Is any extra isolation required when using Analog output to dirve higer power circuitry?

I want to connect the analog output of a DAQcard-1200 to a current amplifier to give 5A and a tranformer to give approx 200V, is any extra isolation required to protect the card?
0 Kudos
Message 1 of 4
(2,807 Views)
Hi Swifty,

If your current amp requires less of everything the teh DAQ-1200 supplies, you should not have to add any thing other than maybe a fuse, just to be sure.

Make sure you look at all of the spec's. Including peak and max values.

Ben
Retired Senior Automation Systems Architect with Data Science Automation LabVIEW Champion Knight of NI and Prepper LinkedIn Profile YouTube Channel
0 Kudos
Message 2 of 4
(2,807 Views)
"Ben" wrote in message
news:506500000005000000408F0000-1027480788000@exchange.ni.com...
> Hi Swifty,
>
> If your current amp requires less of everything the teh DAQ-1200
> supplies, you should not have to add any thing other than maybe a
> fuse, just to be sure.
>
> Make sure you look at all of the spec's. Including peak and max
> values.
>
> Ben

Although it is indeed not absolutely required it is a very bad idea not to
use galvanic
isolation between your DAQ card and a 200 V output amp unless the amp has
already galvanic isolation built in. If the amp breaks and sources its full
voltage to the
DAQ card you not only get a destroyed DAQ card AND computer but possible
claims for injury of the operator who happened to sit at the computer at the
time

the amp blew up.

Rolf
0 Kudos
Message 3 of 4
(2,807 Views)
Wonderful point Rolf!

Swifty, be sure to check as Rolf has indicated.

Thanks again Rolf.

Ben
Retired Senior Automation Systems Architect with Data Science Automation LabVIEW Champion Knight of NI and Prepper LinkedIn Profile YouTube Channel
0 Kudos