Interlock protection with NI 9482

Solved!
Go to solution

Hello,

 

I just want to double check that the following scenario works as I plan.

I have a cDAQ module NI 9482 (link). I would like to use one channel to protect externally my Pfeiffer QMG 220 spectrometer (i measure separatelly the vacuum level at the chamber, in case of too high pressure the LV code must disable the filament). I also have a 24V DC PSU.

I have attached the manual of the QMS, it describes the function of the "EXTERNAL_PROTECTION", page 14. and page 26. with pinouts.

I would connect the components in the following way. Is this config OK?

Thanks!

 

schemeit-project_QMS.png

0 Kudos
Message 1 of 5
(3,974 Views)
Solution
Accepted by topic author Blokk
If so it should work fine. The relay can handle 1.5A at 30V and the external protection input only draws 1.9mA.

Mike

Certified Professional Instructor
Certified LabVIEW Architect
LabVIEW Champion

"... after all, He's not a tame lion..."

For help with grief and grieving.
Message 2 of 5
(3,956 Views)

Thanks!

0 Kudos
Message 3 of 5
(3,952 Views)

Just an important warning for future readers using QMG 220 mass spectrometer!

 

The spectrometer has the electronical control part called "QME 220". Optionally you can purchase an I/O module called IO220. Be aware, if you want to use the "External protection" function, but you do not have the IO module, then you need to refer to page 15 and page 29 in the manual! So you need to use a switch, a floating contact to use the "External protection" mode. This means, DO NOT connect 24V DC to pins 4 and 8! (I did so, and the control unit got damaged, cannot turn on the filament any more, sending for repairment... 😞 )

 

If you have the optional module IO220, then you have to refer to page 16 and page 32/33. In this case you have a D-SUB 25 connector, and you do need to use a 0 / 24V DC signal to use the "External protection" mode!

 

Also be aware, the manual looks like contain wrong  D-sub (9-pin, female) pin numbering! See manual page 29: the shown pinout of the female D-sub connector looks like wrong. On the real device the numbers from 6 to 9 are in reversed order (as for any common D-sub connector), check this on your mass spectrometer unit, if in doubt! I have contacted Pfeiffer and waiting for answer about this issue, so whether I should refer to their manual when wiring or what I see with my eyes 😄 Not risking another unit to get damaged... 🙂

Message 4 of 5
(3,673 Views)
Got the response from Pfeiffer:
"At the manual the numbering of X3, D-Sub, 9-poles, is wrong, as you already assume.
 
Wrong at manual:
image001.png 
 
Right numbering is:
image016.png 
 
The connecting diagram at the manual is right.
 

image017.png 

0 Kudos
Message 5 of 5
(3,641 Views)