In Labview 8.6 and MAX, can I edit a scale for a voltage read channel in a task once the application has been built and installed.

Solved!
Go to solution

In Labview 8.6 and MAX, can I edit a scale for a voltage read channel in a task once the application has been built and installed on a second PC. In fact, where is the scale info held?

0 Kudos
Message 1 of 4
(2,433 Views)

Hello,

 

If you are looking to create a scale that can be edited from within a built application I would like to point you two one of these two KnowledgeBase articles.  These articles outline the methods for creating a custom or programmatically set scale.  

 

http://digital.ni.com/public.nsf/allkb/12288DEB3C6A185B862572A70043C353

 

http://digital.ni.com/public.nsf/allkb/022EF0C2D86CD7F28625704C007B1E5D

 

Regards,

 

Marcus 

Marcus M.
PXI Product Support Engineer
National Instruments
0 Kudos
Message 2 of 4
(2,410 Views)

Thanks for the reply. I have been able to read most of referenced docs. But being old, befuddled and generally thick, I am still unclear as to one aspect.

Once I have the application up and running on a separate computer, is it possible to edit the scale on that computer,. presumably this would be done with MAX. ( This is in Labview 8.6, on DELL PC's, Windows XP, with a PCI 6221 (D37))

 

 Or must I rebuild the application.

 

(To date, I have only built applications where the user inputs calibration constants. These can then be filed and re-read from ad hoc text files. This has suited since they had very much reconfigurable inputs. The application to hand now has a completely fixed configuration for its transducer hardware,  It seems to me it would be cleaner/simpler/safer to deal with only "user units" in the program itself. But the scales would need a facility for re-calibration periodically.)

 

0 Kudos
Message 3 of 4
(2,407 Views)
Solution
Accepted by topic author KGM

Hi,

 

Once you have built your VI and compiled it into an executable and deployed it on a different machine it essentially becomes a read-only file, meaning that you will be unable to edit the scale within this deployed application.  

 

Regards,

 

Marcus 

Marcus M.
PXI Product Support Engineer
National Instruments
0 Kudos
Message 4 of 4