samantham wrote:
Hello,When using the PCI-6024E and the PCI-6259, are you using the same code and task configurations. I am specifically interested to clarify that both are using the same scale and mode (single ended or differential). I am unsure why you are limited to 29 psi on the 6259, which is related to 2.16 V as an analog input and 8.7 mA for the sensor output. Can you verify that the signal is in fact higher than this at the time you are measuring 29 psi?
samantham wrote:
Hello,Lets try testing the 6259 by creating an analog output 5 V and wiring it into an analog input channel. Another good test will be to send a 0-5 V sine wave into an analog input to check its response. All this testing in independent of the code and configurations you are already using.Then, I would suggest testing an analog input with your code but where the signal is coming from an analog output from the 6259 rather than the sensor. I understand your code is intended to test current. So this last test would be without the 249 ohm resistor.Let me know if the 6259 is working with these tests.