I have been using the driver given by national instruments for the read measurement on the agilent 4980. The measurements are made fine but after the measurement is done, I can no longer control the physical LCR meter. What changes might I make to the program to return control to the physical LCR meter?

0 Kudos
Message 1 of 6
(3,055 Views)

I know on some instruments there is a local/remote button that needs to be pushed to get back to local control.

0 Kudos
Message 2 of 6
(3,042 Views)

Is there some way to do this from labview?

0 Kudos
Message 3 of 6
(3,038 Views)
Solution
Accepted by topic author hmar14

I don't have any agilent instruments, so I have no idea.  You would need to find and go through the communication manual for the 4980 and see if there is a command.  You should be able to get it from the manufacturers site if you are lucky.

 

Worst case there may be a reset command and then you would have to stop communicating with it, otherwise it would probably just go back into a remote status.

0 Kudos
Message 4 of 6
(3,036 Views)

Found the button. Thanks!

0 Kudos
Message 5 of 6
(2,976 Views)
If you want to do it progamatically as you asked, you just have to use the VISA GPIB Control REN function. You can use it to issue a GTL command.