LabVIEW

cancel
Showing results for 
Search instead for 
Did you mean: 

How to generate response file for NI 9229 device

My target is to establish a response file for the 9229 device. Response parameters typically consist of a sensitivity or "gain", measured at some specified frequency, and a set of poles and zeroes for the transfer function that expresses the frequency response of the NI-9229. 

 

The 9229 is a 24-bit ADC, and note that there is only one input range, +/- 60 V, on the 9229.  So, per that equation, (120 V)/([2^24] - 1) = ~7.2 uV we can calculate the sensitivity. However, for poles and zeros, I can't find any information in the manual about them. Kindly, how to solve this problem.

0 Kudos
Message 1 of 4
(1,616 Views)

Hi Emad,

 


@Emad_NRIAG wrote:

Response parameters typically consist of a sensitivity or "gain", measured at some specified frequency, and a set of poles and zeroes for the transfer function that expresses the frequency response of the NI-9229. 

 

However, for poles and zeros, I can't find any information in the manual about them. Kindly, how to solve this problem.


Well, in the datasheet for the NI9229 you will find explicit information on passband and stopband characteristics!

There's a whole chapter on filtering behaviour and a lot of numbers in the specs section…

 

Do you need more information?

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 2 of 4
(1,600 Views)

Thanks, Mr. GerdW for your cooperation. Yes, I need more information if you have. Or how to use these information to calculate poles and zeroes (transfer function) for NI-9229 device?

 

0 Kudos
Message 3 of 4
(1,591 Views)

Hi Emad,

 


@Emad_NRIAG wrote:

how to use these information to calculate poles and zeroes (transfer function) for NI-9229 device?


It's been a while since I had to calculate transfer functions using poles/zeros - maybe 20 years ago.

So I guess you have some books explaining those calculations and you got trained to apply them!

 

The datasheet mentions

  • a passband frequency of 0.453*fs with a flatness of <= 0.1dB
  • a stopband frequency of 0.547*fs with rejection of >=100dB

There surely are formulas to convert those numbers into poles/zeros in the imaginary plane…

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 4 of 4
(1,573 Views)