I have a 9189 cDAQ chassis and several NI 9231 cDAQ modules. 

 

How many AI tasks with different sample rates can I have.

 

Thanks.

 

Ian 

0 Kudos
Message 1 of 6
(2,507 Views)
Solution
Accepted by TulaRen

3


"Should be" isn't "Is" -Jay
0 Kudos
Message 2 of 6
(2,492 Views)

Thanks. That's what I remembered, but I just cannot find anything reference to this from ni.com.

 

Ian

0 Kudos
Message 3 of 6
(2,488 Views)

Its down deep in a cDAQ whitepaper


"Should be" isn't "Is" -Jay
0 Kudos
Message 4 of 6
(2,478 Views)

Hello Ian,

 

In this case you can actually only run 2 hardware timed analog input tasks with different sample rates because the 9231 is a DSA module. This is because the cDAQ 9189 is a Gen II chassis and Gen II chassis support a maximum of 2 synchronization pulses. This is called out here.

 

Regards,

Conrad

C Series IO R&D

Message 5 of 6
(2,470 Views)

Thank you for the correction and link


"Should be" isn't "Is" -Jay
0 Kudos
Message 6 of 6