Right now, I can't check the version you just downloaded but I have run this VI before in older versions of the driver and had no problems. What exactly do you mean by 'no response'. Does the voltage come back incorrectly, is there a LabVIEW error generated, or does the instrument indicate an error condition? Can you open/close relays on the 34904? Can you take a voltage reading with the instrument's front panel? Does your model of the 34970 actually have the DMM installed?
The first thing I would do would be to take the measurement manually. If you verify that the instrument does have the DMM installed, try the 34901 in a different slot. If you suspect the 34901, these are steps that Agilent tech support would have you do before returning the card.