08-27-2014 06:41 PM
I recently installed Labview 2014 along with all the other things I need (such as NI-motion, PXI stuff, etc.) and the install went successfully.
I currently have a PXI-1073. which contains a PXI-5152 and the TEGAM4040B instrumentation amplifier, and I have a MID-7602/7604 stepper motor driver that is all connected to the computer to control an experiment where I have to change the angle of an optic and blast it with a laser/collect amplified signals in sequence, etc.
When I reinstalled all the peripherals work fine except that I am now getting the following error whenever I run the sample program that TEGAM gave me to run the amplifier (ie TEGAM4040 - Control [Demo].vi). This .vi definitely worked before I installed Labview 2014. How do I go about troubleshooting this/what is wrong?
"Error 7 occurred at Call Library Function Node in Tegam4040 -- Open4040.vi->Tegam4040.lvlib:Tegam4040 - Initialize.vi->Tegam4040 - Control [Demo].vi
Possible reason(s):
LabVIEW: File not found. The file might be in a different location or deleted. Use the command prompt or the file explorer to verify that the path is correct.
=========================
NI-488: Nonexistent GPIB interface."
Help me LabVIEWwan Kenobi, you're my only hope... I have attached the TEGAM program .vi for convenience.
Solved! Go to Solution.
08-28-2014 04:44 PM
Hi,
This error generally occurs if you provide a path that does not exist for dynamic DLL loading. Do you get the same error when loading other DLLs, or just with this DLL particularly?
08-28-2014 05:01 PM
I contacted the TEGAM engineer and we were able to get it sorted out with a new updated install package. However, thank you very much for your help. 🙂
08-29-2014 08:17 AM
I am glad you were able to resolve the issue! Thank you for letting us know!