Dear all,

 

I use Keithley 2700 with serial communication. If i loose communication with instrument after hours of data acquisition I must restart computer.

 

What does not work for me:

- reconfiguring communication runtime,

- reset keithley (power off/on),

- close and start labview application,

- change of communication cable from serial to USB cable

- software updates: NI 488.2 16.0 to 19.5, NI Serial 17.5 to 19.5, Installed: Keithley I/O Layer version C10

 

This issue with communication still occurs even after NI software updates.

 

I get the following error:

'Error -1073807339 occurred at VISA Read in Keithley 27XX.lvlib:Error Query.vi->Keithley 27XX.lvlib:Error Query (Multiple).vi->Keithley 27XX.lvlib:Configure Scan.vi->SUB = KEITHLEY INIT.vi->MAIN = test.vi'

 

Could anyone please share ideas to fix this issue? What can this issue be related to?


Kind regards,

Balázs

 

 

0 Kudos
Message 1 of 5
(2,838 Views)

Is it possible, that data transfer buffer is overloaded? 

0 Kudos
Message 2 of 5
(2,769 Views)

It would help to see your code.  If possible reduce the code to a simple test case.  That way we can spot potential errors.

 

Without seeing the code, I can say that I've had issues with USB-serial converters and also with USB power settings (Windows selective suspend) dropping out and causing problem like yours.  Power settings help, see here - https://www.windowscentral.com/how-prevent-windows-10-turning-usb-devices

 

When selective suspend happens the software still has an open handle to the instrument but its no longer valid, and only a restart fixes things.  

 

For long term measurements use LAN or GPIB and avoid USB.  I've gone back to PCIe-GPIB cards over USB-GPIB adaptors and those fixed the problems I was seeing.

 

Hope that helps.

Craig

0 Kudos
Message 3 of 5
(2,759 Views)

Hi Craig,

 

You can find my VI as attachement.

 

Kind regards,

Balázs

0 Kudos
Message 4 of 5
(2,739 Views)

If any error triggers a reset then your code needs to a shift register so that the VISA resource ID gets passed back to the read inside while loop catches the new ID.  Wires on one right side of a loop only pass values back to the left side of a loop if you use a shift register. Like so ..

 

 Keithley2700scan_cds.png

 

This still might not be enough to solve your problem, but at least if its possible to restart communications then there's a chance it would work.

 

Craig