Don't understand Error code

I was reading a specific channel on a 6025E using the test of the card, i had this error code ;Possible reasons:
NI-DAQ LV: A channel, port, or counter is out of range for the device type or device configuration; or the combination of channels is not allowed; or the scan order must be reversed (0 last).

with a multimeter i can read 1,2V dc at this point (the right voltage). Could you help?
Caroline Paulin
R&D technician
Astroflex inc
Canada
0 Kudos
Message 1 of 2
(2,669 Views)
What comes to mind first is that your reading is being done in differential mode and you have picked a channel that is out of range. remember that in diff mode the channels are in blocks of 8.

e.g.

0:7
16:23
32:39
if you are in diff mode and try to read channel 9 for example you will get that error.

Dan