DigitalMultiMeter at serial port

Hi all,

i´m searching a solution to read out some data from a DigitalMultiMeter
"Voltcraft M-3850D" at the serialport (1200bit/s,N,7,2). I have tried it
with the standart serial vi´s, but without success. The DMM waits for a "D"
command and answers with a series of 14bytes. I also have tried it with the
IN & OUTport vi´s, only with a little success, because there ist no
syncronisation between pc and DMM.

Has perhaps somebody one solution for me?
Perhaps has someone this DMM with LV on active service.

--
MfG / best regards
M.Waschk mailto:waschk@mavd.de
0 Kudos
Message 1 of 3
(2,888 Views)
There was a similar question yesterday. Check out my answer at http://exchange.ni.com/servlet/ProcessRequest?RHIVEID=101&RPAGEID=135&HOID=50650000000800000000230000&UCATEGORY_0=_49_%24_6_&UCATEGORY_S=0. I would also add that you should always configure the serial port inside of the VI before doing a read or write.
0 Kudos
Message 2 of 3
(2,888 Views)
Again hello...

we have found the problem. It is necessary to set the handshake. In our case
we write a "9" to the base address 3F8 +4.
If anyone want the vi, send me a e-mail.
--
MfG / best regards
M.Waschk mailto:waschk@mavd.de