I had a question again: how do I set the address of this flow meter ?I knew it's a silly question but I have no idea how to get it work. for Adam6066 that I used MODbus TCP, I can read and write the registers very well. but go to the RTU, I have no idea how to set the communication.
so, again, my set up is I have a Flow meter which is from Bronkhorst, i connected it with my laptop, via using a cable with one end is RSS232 and another end is RJ45. I have to use MODbus RTU to bulid the communication. However I dont know how to set the address to start the communication. is anybody can give me an example in this application? many thanks
07-24-2009 03:53 AM
hello echomice, there are two ways to set the address of your bronkhorst sensors/controllers.
1- using a software---> in your case will not help you becuase you still dont get communication with your device.
2- Hardware using the NODE ADDRESS switch on the side of the instrument --->The MSD is the high part of the decimal number and the LSD the low part. For instance address 14 means MSD on 1 and LSD on 4.