I am working on a project in which I have to output a square wave signal through USB-4431 and apply it to a circuit. Then I have to input the response signal from the circuit back to the USB-4431 and plot it against the output signal. The program attached to this query plots the two signals but there is a delay in both of the signals.
How can I get the output response of the circuit and the output square signal at the same time ?
(1,385 Views)
Few points why you don't get the right response,
  1. DSA devices are not an appropriate instrument to generate a square wave, DSA devices are meant to generate clean sine waves (you would have observed that the generated square wave is not really a square but a combination of sines)
  2. You need to synchronize both AO and AI - refer to the available techniques here https://www.ni.com/en-us/support/documentation/supplemental/10/dynamic-signal-acquisition--dsa--sync...
    • You can use sample clock synchronization to keep it simple
  3. Please note that there is always a delay associated with the digital filters at the input/output of delta-sigma ADC/DAC, this may vary and you need to compensate for those
Santhosh
Soliton Technologies
(1,347 Views)
Hi Yum,
some hours ago you asked the same question with a USB-6003 in mind
The same techniques apply for the solution: use the same sample clock and/or a start trigger for both tasks.
LabVIEW comes with a huge example library, which contains examples on DAQmx usage…