faster data transfer on ethernet
Hi friends,
I am using VISA to control instrument through ethernet tcp/ip protocol.
the application is written in VB6 with the help of  measurment studio controls
i need to read  large matrix data , element by element ,since the insrument has no feature of bulk transfer
it is taking too long time to read the entire matrix .
the thing i observed was there is a 100mSec Delay between Visa Write and Visa Read ,(Visa read responses back after 100msec)
 this i observed with the help of  NI 488.2 SPY.
(this problem was not observed when controlling without the Visa )
is there any explaination why this 100msec delay is coming for each read , which amounts to enormous time for 30,000 elemenst read.
shridhar
(3,352 Views)
Accepted by topic author shridharjosh
Most of the defaults for VISA are set to optimize for large bulk transfers, since that is the norm for TCP/IP instruments. If you are transferring small amounts of data at a time there are several properties/attributes that you can change to get much better performance. Open up the VISA Help and take a look at the following attributes:
VI_ATTR_SUPPRESS_END_EN
VI_ATTR_TCPIP_NODELAY
VI_ATTR_TERMCHAR
More than likely you will just need to change the first two, because your instrument probably does not support any type of termination character.
-Josh
(3,337 Views)
