dynamic time delay between wright and read.

Hi All

 

I'm woring on expanding my instrument sriver for my Agilent DSO-X4043A Oscilloscope.

I have made a VI that sets up a VPP measument and it works fine. Until I set the aqusition type to "Average" and ask it to do this over 128 measurements.

 

I suspect that this is due to the fact that it takes to long to do this averaging. So that I try to read before the data is ready.

Since I'm running this over ethernet, I dont have the otion of monitoring the amount of data ready to be read.

 

My question is therefore: Is there a way to determin when the scope is ready to "unload" the data?

 

I know I could set up a time delay, but I'm duing quite a lot of measurements. So I would like to have this VI running as fast as posible.

 

Help is much areciated

 

Regards Rasmus

0 Kudos
Message 1 of 4
(4,069 Views)
Have you tried an OPC? query? That stands for Operation Complete.
0 Kudos
Message 2 of 4
(4,067 Views)

Hi Dennis

 

Yes, I just tried that without luck. 😞

 

I get the -1073807339 timeout error after my read command and when I check the scopes error log afterwards I get a "Instrument reports:
-221,"Settings conflict"" massage.

 

I can however not anything in the instrument manual about conflicts with the aqusition type...

 

Regards Rasmus

0 Kudos
Message 3 of 4
(4,061 Views)
What does the manual say about the status byte? Perhaps you can read that.
0 Kudos
Message 4 of 4
(4,049 Views)