09-18-2005 01:14 PM
I am hopeful that you GPIB gurus can help me.
I am new to GPIB. I am trying to interface a Tektronix 2710 Spectrum Analyzer to a National Instruments GPIB-232CT-A.
***Please see the attached Word file to see Software set up and Win32 Interactive Control information.***
My computer is a 333 MHz Windows 98se PII and I have the GPIB-232CT-A interconnected to the COM2 port at 9600 baud with a 9 pin RS-232 null modem cable and am using a new NI 2.1 meter X2 GPIB interconnect cable between the TEK/2710 and the GPIB-232CT-A. I am using the most recent NI.488.2 GPIB drivers (2.7) available for the GPIB-232CT-A. NI's diagnostics programs respond with the interface working.
The GPIB interface on my Tektronix 2710 Spectrum Analyzer doesn't give many options. I have it configured for Status: "Online," Address: "1", Power On SRQ: "OFF," EOI/LF Mode: "EOI", and Talk Only Mode: "OFF." When I run the 2710's internal GPIB diagnostics program, it responds with "Success."
One would think that with these settings, it would work. But it doesn't.
Do I have a bad GPIB board or controller?
If you have any ideas on how to test the interface and get it working, I would very much appreciate your help.
Brian
09-19-2005 03:17 PM
09-19-2005 03:39 PM
Hi Robert,
This is the first time that I have tried to use this controller. It is also the first time that I have tried to interface with my spectrum analyzer. Unfortunately, at the moment, I don't have any other GPIB controllers or and other GPIB devices.
At least to me, it is beginning to look like I have a hardware problem.
I wanted to use the GPIB-232CT-A so I could use a laptop in the field.
NDAC seems to indicate that the device is on line. I had originally configured the analyzer on GPIB address 1. I switched it to GPIB address 0 just to see what it would do. It actually almost works on address 0, and does practically nothing on address1. Go figure.
Are there any easy definitive ways to test the hardware other than to substitute other units?
Brian
09-20-2005 02:18 PM
09-23-2005 09:00 AM
Robert,
I thought you would be interested in knowing our test results.
Apparently, the GPIB interface in my Tektronix 2710 Spectrum Analzyer is not working properly.
Hi, Brian --
Well, I got the adapter and cables this evening, and just got through trying
them on the Win98SE box in the basement lab. It looks like you are in good
shape vis-a-vis the GPIB hardware; it all worked perfectly.
I was able to talk to the 8566, the 494AP, and the
2430A. All of the settings are at their defaults; I didn't have to change
anything.
There seems to be a long startup period associated with sending or receiving
traffic (not sure which). My phase-noise utility does a lot of
back-and-forth communicating with the SA, and it took about 10 minutes to
run on the GPIB232 versus 2 minutes to take the same series of traces via my
PCI-GPIB. But I never saw any errors or flaky behavior.
Not sure what to recommend at this point. 😞 , but I
guess you're running out of stuff to try!