Visa Interactive Control

Solved!
Go to solution

 

Hello,
I am trying to connect an Almemo 2890-9 with LabVIEW. The device is connected via a ZA1919-DKU, which is an RS232 to USB adapter. I have tried to make the connection using NI MAX through the VISA Interactive Control, but I get an error (see image). I have checked the Baud Rate/Data Bits/Stop Bits/Parity settings (9600/8/1/none) and everything is correct. I have also increased the timeout. Alternatively, I tried using the free Almemo software (Almemo Control v6), and it detected the device perfectly. What could be the problem? Thank you very much.

Jousepa_0-1754044564221.png

0 Kudos
Message 1 of 2
(691 Views)
Solution
Accepted by topic author Jousepa

Is Port Connected is only valid for NI serial boards. I would not worry about that one.

 

Assuming you have the UART settings correct (Baud Rate, etc.), I would be double checking the query command and the termination character.