I saw the VIs you attached. (Sorry but I could not run them because of different IO environment.)
The VIs seem like include some power-supply control routines using IviDCPwr class driver and direct GPIB Write. And I found that the VIs - "Power Supply ON or OFF" send GPIB commands for the supply. Is the power supply that appeared in the "Power Supply ON or OFF" and "Current Consumption Test 1.01" the same instrument? If so, you should control VOLTAGE/CURRENT/OUTPUT of the power supply through the IviDCPwr using Configure Voltage Level, Configure Current Limit, and Configure Output Enabled VIs, because your code already has IVI dependent parts in "Current Consumption Test 1.01".
Also, what is the relationship between the VIs you attached and Agilent 34405 DMM? If you want to measure the consumption current values by DMM, you should use IviDmm class driver instead of IviDCPwr. If using IviDCPwr for measurement processing, it means that you measure the current or voltage by using power supply's embedded measurement functions.
09-19-2006 12:51 AM
Hi Makoto
Your are right! The 3 VIs are work in a TestStand sequence, the step is on,measure and off for a same instrument.
Follow your advice, I will change GPIB write to Configure Voltage Level, Configure Current Limit, and Configure Output Enabled VIs
I measure current by using power supply's embedded measurement functions.