VISA read timeout error--- Keithley 2400

Résolu !
Accéder à la solution

here is initialize.vi. the first half went well, but error pops out from VISA read. please check the attached. thanks. please let me know if you can open it or not.

0 Compliments
Message 11 sur 24
7 980 Visites

Hi,

 

Can you run your code in Highlight execution mode and let me know at which point in your code does the error occur. Is it at the VISA read which is right after the VISA write?. It would be helpful if you could take the screenshot of the error message and post it.

Sev K.
Senior Systems R&D Engineer | Wireless | CLA
National Instruments
0 Compliments
Message 12 sur 24
7 975 Visites

You forgot to make all values default, after the errormessage

 

go to edit/make current values default and save the vi. I was not really interested in the vi but in the status of the frontpanel controls and the errormessage.

greetings from the Netherlands
0 Compliments
Message 13 sur 24
7 969 Visites

Hi Sev and Albert,

    It happens right after the VISA write function, i have attached the screenshot for the initialize.vi and highlight the errors. The error message that comes out from the front panel is "VISA Read in Keithley 24xx. lvlib: Error query.vi->Keithley 24xx.lvlib: Close.vi->Keithley 24xx.lvlib:Initialize.vi"

Tout télécharger
0 Compliments
Message 14 sur 24
7 957 Visites
Solution
Accepté par l'auteur du sujet Michigan state

Let's get this error fixed by doing the instructions below:

 

Right click on the output of the VISA Read and select 'Create Indicator' (the pink wire that is already there)

Run the VI - making sure there is something in the indicator you just created.

Go to the Edit menu and select 'Make Current Values Default' (you were told to do this once before)

Save the VI

Attach the VI.

 

The images are not worth much and doing the loopback test is less than useful. Since you are able to execute the command in MAX, you should get it to work in LabVIEW.

 

I suspect the instrument is set for a different termination character than the Line Feed you are currently using. This is something that should be explained in the instrument manual.

Message 15 sur 24
7 949 Visites

Dennis,

    YOU ARE CORRECT.....

    The instrument is set for a different termination char. i changed it from <CR> to <CR+LF>. Problem is solved.

    Thank you very much for your advice and also thank everyone who contributed their time to help me look into the problem.

    You guys are awesome!Smiley très heureux

0 Compliments
Message 16 sur 24
7 918 Visites

Then you should mark my answer as the solution. You can change this by clicking the 'Options' link.

0 Compliments
Message 17 sur 24
7 910 Visites

Hello, may I ask where did you change CR to CR+LF??

0 Compliments
Message 18 sur 24
7 329 Visites
On the keithley machine
greetings from the Netherlands
0 Compliments
Message 19 sur 24
7 319 Visites

I am not sure the 2602 has it or not. 

0 Compliments
Message 20 sur 24
7 310 Visites