VISA Serial read error

Solved!
Go to solution

I´m trying to read data from my OHM60 Heratherm oven with a RS-232 cable and Visa Test Panel /Visa interactive control. 

I´ve followed the OHM60 document and enter the wanted parameters:

— 57600 bits per second
— 8 data bits
— 1 stop bit
— No Parity

Also entered the given commands from the document, (?:3010:00::c2) in order to read the temperature. But when sending Read/Query i get this error:

"Read Operation
Error

VISA: (Hex 0xBFFF0015) Timeout expired before operation completed."

 

I´ve looked around this forum and googled for many hours but can´t find a solution.

Please help.

/D

0 Kudos
Message 1 of 17
(3,195 Views)

Did you finish with a crlf or other endcharacter specified in the manual?

greetings from the Netherlands
Message 2 of 17
(3,186 Views)

My first thought is that you need to add a Line Feed to your command in order for the instrument to know the message is complete.

 

Otherwise, I recommend you post your code and we may be able to spot something that is wrong.


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
Message 3 of 17
(3,184 Views)

Thanks for the fast response!
I did not really write any code just wrote the simple command line as in the picture attached. What kind of Line Feed should I add ?

Sorry if I am a big ignorant, I am quit new to this. Thanks in advance.

I attached the page from the manual with the command syntax and so.

/D

Download All
0 Kudos
Message 4 of 17
(3,179 Views)

Thanks for the fast respons!

No I do not think so, in the  Visa Test Panel I filled in the parameters and pressed the "Input/Output" tap, then enter my command which is the ?:3010:00::c2 as I mentioned and then pressed Query and/or Read. Which gave the error.

See picture "Visa test panel.png"

/D

 

Download All
0 Kudos
Message 5 of 17
(3,176 Views)

Look at the manual on page 5-13.  Try adding the <CR> character to the end of your query.  It should work with the first line.  If not, try the other.

?:3010:00::c2\r
?:3010:00::c2\n
Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
Message 6 of 17
(3,172 Views)

I have tried "?:3010:00::c2<CR>" but not the /r and /n, I will try it and respond when I can. Thanks!

/D

0 Kudos
Message 7 of 17
(3,170 Views)

@wasa1 wrote:

I have tried "?:3010:00::c2<CR>" but not the /r and /n, I will try it and respond when I can. Thanks!

/D


The "\n" when in "\ Codes" display represents a <CR><LF>.  Likewise, "\r" will represent a <CR>.  If you are in a normal/ASCII only display, this will not work for you.


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
Message 8 of 17
(3,165 Views)

Sorry but when trying to use \r and \n as you mentioned it does not work. Any ideas?

See attached picture: Visa test panel1.PNG

/D

0 Kudos
Message 9 of 17
(3,135 Views)

Just to make sure, you are pressing the QUERY button and not the WRITE button?

Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
Message 10 of 17
(3,127 Views)