Unexplained delay in Python - PyVISA - USB - GPIB - Keithley 2401 Stack
I am trying to run an IV sweep (sweep voltage, measure current) on a Keithley 2401. I am using Python 3.10, PyVISA, and controlling a Keithley 2401 via USB to GPIB.
This used to "run fine;" then I started seeing occasional random ~17-second delays (in the middle of measurements); now nearly every measurement has an identical ~16.7 second delay.
I am not posting code in order to avoid tripping CloudFlare warnings. I will edit this post if it succeeds.
(750 Views)
Accepted by topic author GoMrPickles
    # Query the identification string
    idn = ksm.query("*IDN?")
    print(f"Successfully communicated with: {idn}")
When trying code as simple as this (where `ksm` is a class corresponding to a Keithley 2401 SMU), we would see random ~17 second delays. Eventually, these delays stopped being random, and occurred with nearly every command we'd send via PyVISA to the Keithley.
I eventually tracked this down to the TNT4882 ASIC in the GPIB-USB adapter. The NI drivers (2025) were settings the bus to high speed; it needed to be normal. Downgrading to the 2020 drivers did not fix it. Setting the bus timing to 2 us in the NI MAX control panel didn't fix it.
Eventually, I (with help) looked at the Linux GPIB drivers, and found the file that talks to this ASIC:
https://github.com/greyltc/linux-gpib/blob/master/drivers/gpib/include/tnt4882_registers.h
What did eventually work was explicitly setting the hex register to the correct value, in Python, every time I instantiate the `ksm` class.
# Force T1 Delay to 2us (Value 1 = Normal/Slow)
# This ensures the ASIC doesn't 'outrun' the Keithley handshake
k.visalib.set_attribute(k.session, 0x3FFF0184, 1)
In the VISA and NI-488.2 standards, the T1 Delay (VI_ATTR_GPIB_T1_DELAY, Attribute 0x3FFF0184) usually maps to specific timing states:
  • 1: Normal ($2\mu s$)
  • 2: High Speed ($500ns$)
  • 3: Ultra High Speed ($350ns$)
  • 0: Default/Unconfigured
The Default State (0): In this state, the NI-VISA 2025 layer may allow the hardware to use its fastest possible timing (350ns), which triggers the 16.7-second hardware timer when the Keithley 2401 cannot keep up.
By setting the T1 delay within my KSM class in Python, I no longer see this delay.
(745 Views)
Hello.
Where did it say that 0x3FFF0184 indicates a T1 Delay?
In typical VISA references, this hexadecimal number is assigned to VI_ATTR_GPIB_UNADDR_EN.
(556 Views)
You're right, it is... and I haven't found a specific reference which says that 0x3FFF0184 corresponds to VI_ATTR_GPIB_T1_DELAY. I actually can't find any docs for "VI_ATTR_GPIB_T1_DELAY" now. Either way, setting that hex value to 1 solved a problem that took down one of our tools for a month.
Some further analysis:
I can set 0x3FFF0184 to any value between 0 and 16960; if I try a higher value than 16960, I get a silent error and it falls back to 16960. That should not be possible, if it's a Boolean:
https://www.ni.com/docs/en-US/bundle/ni-visa-api-ref/page/ni-visa-api-ref/vi_attr_gpib_unaddr_en.htm...
Setting the bus timing in NI MAX was not sufficient to fix the error; Google searches now return mostly my own post here. If the error is unrelated to that bus timing, that explains that part.
NI I/O Trace shows that 0x3FFF0184 is VI_ATTR_GPIB_UNADDR_EN. Enabling it forces the controller to unaddress after each transfer. Without it, the Keithley 2401 can leave the bus in a state that causes the TNT ASIC to hit its 2^24 µs timeout (~16.7 s) during the next write. I attached an I/O trace.
Frankly not my area of expertise...
(528 Views)
Hello.
Thank you for your reply.
It seems to be some kind of hidden attribute. Or it may just be working by chance due to a bug.
The GPIB settings made in NI-MAX are saved in C:\ProgramData\National Instruments\NI-488.2\gpib.ini, but these settings are probably not used when using GPIB from VISA.
The settings here are probably applied when calling ni4882.dll or gpib32.dll directly without using VISA.
(483 Views)