Timeout in ibrd long before timeout has expired.
As you can see it set a timeout of T1000s and then sent a command to a device followed by an ibrd.
however the ibrd terminated pretty much immediately even though the device was still processing the *CAL? command
What could be going on?
David
(3,472 Views)
Looks like the instrument responded, possibly with a \r or \n or NULL character.  But without seeing your code or knowing the instrument in question it will be hard to diagnose the problem. A few questions that might lead you to a solution..
Are your termination characters properly setup?
Is the value T1000s defined in your code?  (I believe its in a header file somewhere, otherwise you can supply an float 1000.0)
If you read the status does it give you an error?
Hope that helps.
Craig
(3,440 Views)
T1000s is defined in NIGLOBAL.bas as the value 17 which gives a 1000 second timeout.
> Global Const T1000s = 17 ' Timeout of 1000 s (maximum)
The instrument is a Tektronix TDS3000B 'scope
AFAIK a status of 0xc100 says timeout with no response.
David
(3,433 Views)
VERY interesting - I changed the timeout value from T1000s to TNONE, and that works fine !
The trace did show a value ot 17 being passed to the ibtmo call previously so it's not as if I had been passing an invalid value ...  However when it didn't crash at that point it did consistently timeout after about 15s (the default IIRC)
Application appears not to be inclined to crash randomly anymore hmmmm
David
(3,412 Views)
I'd like to see the chain of read write commands you are making to understand how you use the timeout.  Are you connected by USB, GPIB or LAN?  PC or laptop.  USB can be flaky for long delays.  I had a DMM which ran the following
ibtmo(hp3458a, T1000s) ;; Set very long timeout!
ibwrt(hp3458a, "CALL TEST1", 0)  ;; call subprogram measurement setup
ibwrt(hp3458a, "STB?", 0) ;; read status
ibwait(hp3458a, TIMO|RQS) ;; wait until the program is finished reading or timeout
stb_rqs=ibrd(hp3458a, 3) ;; read return value then decide how to proceed
works fine on GPIB with PCI card but fails 4/10 times with GPIB-USB adaptor due to PC USB port issues.
Yours doesn't look like the same issue, but without more code hard to diagnose.
Craig
(3,410 Views)
You may be onto something there - this is connected via a GPIB-USB-HS and it's a laptop.
The code now reads:
   ibask TDS, IbaTMO, CurrentTimeOut   ' find current timeout
   WriteDevice TDS, "*CAL?"
   ibtmo TDS, TNONE                    ' change timeout to None from T1000s
   Msg = Space(20)
   ibrd TDS, Msg                       ' don't use AskDevice or ReadDevice - they use fixed T/O of 3s
   ibtmo TDS, CurrentTimeOut ' reset timeout to default
However - it's odd that the T/O happened after 15s or is that the point at which USB starts giving problems?
David
(3,405 Views)
In case there's any doubt, the code used to read:
ibtmo TDS, T1200s     ' Long timeout
Msg = Space(20)
and yet the ibrd was terminating with a timeout (status 0xC100) after 15 seconds.
Does anyone know why this happens and how to solve it.
David