Tektronix TDS2024C usb issue
    I have a TDS2024C and connected to PC by usb.
    NI Max can detect the device, but when I tried to send command through VISA test panel,  it doesn't work.
    is there any suggestions?
Thanks!
Message 1 of 9
(6,658 Views)
What did you try and what did you expect to work.
Post a screenshot of the detected system if possible.
greetings from the Netherlands
Message 2 of 9
(6,636 Views)
Hi Here is the screen shot.
you can see the Ni Max detected the oscilloscope, but can't send commands to it.
Message 3 of 9
(6,635 Views)
Indeed MAX sees this but MAX only knows when USB is working correctly.
So without hardware to test I only can guess:
I see two USB connectors, I hope you connected to the back
And I see Signal express is on board shouldn't that be disconnected? Or external computer conected?
Message 4 of 9
(6,632 Views)
   Yes, usb is connected to the back of the scope.
    So is there any setting I need to do?
     When I connected the scope to a computer, I installed TekVISA, not sure if it is necessary or it cause the problem. And I restarted computer too. What else I can do to find the problem?
thanks!
Message 5 of 9
(6,629 Views)
Tek visa is probably fighting with max.
Please stay as close as possible to ni software.
Message 6 of 9
(6,624 Views)
Accepted by topic author ForrestGan
For some of these instruments, it prefers the other selection in the Windows Control Panel for USB.  Select the USBTMC which is the USB Test & Measurement Class Device driver.  I have had to use this in the past for some Tektronix instruments.
Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
Message 7 of 9
(6,607 Views)
    the instrument was shown as USB Test & Measurement Class Device in the device manager.
    somehow after I disable and enable the device again, now everything works fine.
     I am still using the usb address shown in the NI Max.
Best
Message 8 of 9
(6,600 Views)
pls send your vi ..how you are communicating to ur instyrument using visa.. or post the snap shots step by step..
Message 9 of 9
(6,001 Views)
