03-16-2019 02:46 AM
I have a TDS2024C and connected to PC by usb.
NI Max can detect the device, but when I tried to send command through VISA test panel, it doesn't work.
is there any suggestions?
Thanks!
Solved! Go to Solution.
03-16-2019 10:56 AM
03-16-2019 11:13 AM
Hi Here is the screen shot.
you can see the Ni Max detected the oscilloscope, but can't send commands to it.
03-16-2019 11:28 AM
Indeed MAX sees this but MAX only knows when USB is working correctly.
So without hardware to test I only can guess:
I see two USB connectors, I hope you connected to the back
And I see Signal express is on board shouldn't that be disconnected? Or external computer conected?
03-16-2019 11:40 AM
Yes, usb is connected to the back of the scope.
So is there any setting I need to do?
When I connected the scope to a computer, I installed TekVISA, not sure if it is necessary or it cause the problem. And I restarted computer too. What else I can do to find the problem?
thanks!
03-16-2019 12:40 PM
03-18-2019 07:39 AM
For some of these instruments, it prefers the other selection in the Windows Control Panel for USB. Select the USBTMC which is the USB Test & Measurement Class Device driver. I have had to use this in the past for some Tektronix instruments.
03-18-2019 12:16 PM
the instrument was shown as USB Test & Measurement Class Device in the device manager.
somehow after I disable and enable the device again, now everything works fine.
I am still using the usb address shown in the NI Max.
Best
03-05-2020 01:48 AM
pls send your vi ..how you are communicating to ur instyrument using visa.. or post the snap shots step by step..