Hi Collin,
Do you mean the software implementation portion aspect? There is only 1 equipment connected to the test computer (master controller) per setup. The GPIB cables used range from 1m to 3m, some setups have 2 shorter cables connected in daisy chain. The equipment is a Delta Castle Logic handler (slave) (IEEE-488.1). The length of the cable recommended by the handler is maximum of 4m. Messages exchanged are base 10 ASCII-coded numbers.
No, random character missing are observed and at different position. There is no specific pattern that i'm able to percieve.
GPIB Settings used:
- Bus timing: 500nsec
- Cable length: Disabled
- System controller: checked
- Enable auto polling: checked
- Send EOI at end of Write: checked
- EOS byte: 0
Eg3.
GP: emm? (emulator mode?)
=> 2 (mode 2 = no emulation)
GP: spcm 255 (set spoll clear mask)
GP: prm 255 (should be sprm 255: spoll read mask)
ERR: Neither 'prm' or '255' a command
Eg4
GP: spcm 255
GP: sprm 255
GP: sqlm 255 (set SRQ line mask)
GP: ts? (should be nts: read the number of test sites)
There is no error messages coming from the test computer/program. As in the example2, trs 1 (missing ,,1). The tester should be giving the 1 result for each site, i.e total 2. However, due to the missing ',,1', the tester gave only 1 result to the handler.
The problem is strange as we were not able to simulate the problem and the occurrence is intermittent.
Apologize for the lengthy post. Thanks.