MAX displays text in hex...
When I communicate with my device with MAX, the scan for instruments displays the string "IPM488", but when I use communicate with instrument, it starts displaying the text in hex ascii value... "IPM488" become "49 50 4D 34 38 38". How can I make it displays the string like when using scan for intruments?
Michael
Message 1 of 4
(3,774 Views)
This is an unusual error. I'm not sure why it is that way? Could you attach a screenshot?
Randy Solomonson
Application Engineer
Message 2 of 4
(3,774 Views)
Here is the screenshot of the NI-488.2 Communicator with string display in hex ascii and NI Spy that displays the string correctly.
Message 3 of 4
(3,774 Views)
The applet scans for any character that seems to be binary data. The alphanumeric, carriage-return and line-feed characters are not considered binary data. However, the null character is considered binary. Thus any string ending with a null character is considered binary.
The solution would be to use the VISA Interactive Control (VISAIC.exe) utility instead.
Randy Solomonson
Applications Engineer
Message 4 of 4
(3,774 Views)