GPIB on CompactRIO
A simple question, but I want confirmation.
Does the GPIB-USB-HS+ (https://www.ni.com/en-us/shop/model/gpib-usb-hs-.html) work on the CompactRIO, or any other third party GPIB to USB controllers?
What are to requirements for it to work on a CompactRIO?
(3,122 Views)
(3,081 Views)
So if I understand correctly, it is not possible to use the USB port for GPIB control, instead I should use the RS232 port in combination with the GPIB-RS232 converter.
If I were to use this converter, can I send the converter the GPIB data that I would send to the GPIB controller, or do I have to send the RS232 protocol for the converter?
Thank you,
Prentice
(3,050 Views)
Accepted by PrenticeK
You are correct Prentice,
The USB will not be recognized by the controller.
The GPIB-RS232 will speak RS232 with the cRIO and GPIB with your device. You will send the commands via RS-232 and they will be forwarded to the GPIB device.
For example: you will send *IDN? via RS-232 and the converter will send *IDN? through the GPIB bus.I believe you would use the S mode, per Which Mode Should I Use During The GPIB-RS232 Communication?
(3,039 Views)
Ah I see. Thank you for your help!
Prentice
(3,024 Views)
