Currently I'm trying to connect to the fluke via LAN - so far being successful achieving connection via IP + Port inside Keysight Connection Expert.
The point is, That for the external script I need the IP + Remote name connection, but I don't know how to acquire one. "inst0" as default, doesn't work. Is there any SCPI command to read the assigned name? Or maybe some PyVisa command? Any advices?
Best regards!
12-06-2022 04:04 AM
Answering my own question: Fluke 8846A doesn't support VXI-11 protocol, which is required to initialize DMM object as instrument class by VISA library. So it's a dead end - rather hardware obstacle.