Error!!! 2 instruments for 1 GPIB slot
Hey Magou,
The simplest way to check your driver version is to open up Measurement & Automation Explorer(MAX) and expand the My System tree. From there navigate down to Software and expand that one as well. This might take a second to populate depending on your system. The next step is to scroll down to NI-488.2 and left click with your mouse on this icon, this will open up the Version, Description, and Path within the center tab.
Thanks and have a great day!
Nicholas K
Applications Engineer
Message 11 of 15
(1,532 Views)
Hi all,
my MAE version is 3.1.1.3004. Instrument 1 is HP 4284A Precision LCR meter. The GPIB cable using is DDK 408JE. I had print screen some of the problems i encountered.
I try to changed by I/O timeout from 1 sec to 1000 sec, but it still the same.
I also try to change the bus timing from 500ns to 2ua. but it still the same.
Everytime i click the "scan for instruments", the main page will shows different instruments.
Instrument 0 - address 5
Instrument 1 - address 17 (HP 4284A meter).
sometime will show only one of them, sometimes will show both or even more.
thanks to advise if any settings is wrong.
Message 12 of 15
(1,513 Views)
The version of MAX probably has nothing to do with your problem. It's the version of NI-488 that could possibly make a difference but I doubt it. The bus timing change was the only hope that I had that this could be a simple fix. I still think that the issue is electrical. You said you tried different cables and a different computer. I would assume that by different computer, you mean you tried a different controller in another pc. Trying the same controller in a different pc is not testing anything at all really. If you did test a different controller, then that leaves a defective instrument. This is a problem that I have seen before with instruments with a bad GPIB interface. Works fine by itself but the bad instrument's control lines have an issue. I asked once before if you could try it with a completely different set of instruments. Are you able to do that?
Message 13 of 15
(1,502 Views)
I unable to try on another set of instruments. I will check the version of NI-488 again.
what do you mean by the bad instrument's control lines have an issue?
Message 14 of 15
(1,493 Views)
A GPIB bus has 8 data lines, 3 handshaking lines, and 5 bus management lines. As mentioned by Steven, before any data is sent/received (with an *IDN?) query, instruments are detected whit the handshaking/management lines. An instrument could have one of the lines stuck and this could prevent another instrument being correctly detected.
Message 15 of 15
(1,485 Views)