Hello jhorvath,
Thank you for contacting National Instruments.
From the information you have provided here, you are correct that you are able to disable the Automatic Serial Polling using the NI-488.2 call ibconfig(bd, ibcautopoll,0) or disabling “Automatic Serial Polling” in Measurement & Automation Explorer.
There is not a VISA call that will turn off the polling because the VISA software layer is for more general commands. I referenced the NI-VISA Programming Manual when looking into your question.
I would recommend that you use the NI-488.2 call or the option in Measurement & Automation Explorer.
I hope this helps! Let me know if there is anything else I can help with or clarify. Have a great day!
Sarah S.
Applications Engineering
National Instruments