Gents,
vi is attached. I have the device set up to run continuously. Also I’m using the current plug and play drivers from NI. I also have it working now as I want. I found the problem to be the input control for the Hz. The default value is 1000. I created a frequency control for the example vi (ag33xxx – Wfm Config Ex.vi). Then set the control value to 500, expecting a frequency of 500 Hz, but was getting 500k Hz. My sampling rate was not fast enough for that frequency resulting in a shot gun type pattern of a sine wave at 500k Hz. Once I divided the input control by 1000 I got the desired result.
Thank you.
03-26-2008 09:47 AM
10-23-2008 04:30 AM
I'm using Labview 6.1 i need agilent 33250a vi but i can use your it's for 8.5 could tou pls save your file for 6.1 version and send it?thnx
10-23-2008 08:27 AM
Hello Greg,
I have attached the LV 6 version of the ag33xxx LV PnP Instrument Driver. Enjoy.
-NathanT