11-22-2010 03:24 PM
I have a USB-5133 that I can use on several computers. I'm trying to put it on another PC (XP Pro). The device is recognized by windows and it proceeds to install drivers. When it's done, under Device Manager -> Data Acquisition Devices, I see "NI USB-5133 Loader". On the other computers it just says "NI USB-5133". The green light on the device is NOT lit, and the device does not show up under Measurement & Automation Explorer.
Any idea why the proper driver isn't loaded? I have NI-SCOPE 3.5.2.
Solved! Go to Solution.
11-23-2010 01:11 PM
Hello Gallamine,
Since the NI USB-5133 Digitizer is a plug-and-play device, try deleting the device in MAX. Then disconnect the device and reconnect it. The computer should register the device in MAX as "NI USB-5133". If that does not work, the MAX database might be corrupted. To fix this, refer to the following link:
http://digital.ni.com/public.nsf/allkb/2C7480E856987FFF862573AE005AB0D9
Lastly, if fixing the database does not solve this issue, try reinstalling the newest version of NI-SCOPE.
Regards,
Roman S
11-23-2010 01:24 PM
Roman,
The device does not show up in MAX. It only shows up as "USB-5133 loader" in the PC's device manager.
I'll try removing the database corruption and report back.
12-05-2020 11:32 AM
Many thanks dude for the tip! You saved me from a 2 hours NI scope reinstall cycle!