Well, may as well extend my streak...
The [c]FP-RTD-122 (3-wire RTD modules) can support resistance measurements up to 4000 Ohms. The [c]FP-RTD-124 (4-wire RTD modules) will support resistance measurements up to 400 Ohms.
If you need to measure larger resistances with FieldPoint/Compact Fieldpoint, you can hook an additional resistor in parallel to the 10-100K resistors to bring the total resistance down to the range that the module supports (4K). This will allow you to measure the parallel equivalent resistance and back-calculate the resistance of the thermistor. Unfortunately, in doing this, you will lose both accuracy and resolution since you are now looking at the parallel resistance. Alternatively, placing a known voltage across the thermistor and
measuring the current through the thermistor would allow you to calculate the resistance, but you may run into ground loop issues with this approach since there currently are no channel to channel isolated Compact FieldPoint modules.
Regards,
Aaron