4461 Analog output performance

Solved!
Go to solution
Solution
Accepted by topic author ON

The THD is around -110dB.  I thought that the spectrum analyzer is loading it adervsaly so I decided to terminate the input of the spectrum analyzer with a 1Meg ohm with 200pF cap in parallel.  After I did this the results improved dramatically.  not the second harmonic is down around -100dB.  See Attachment.  I believe I can get better results by terminating the input of the ADC with 1Meg ohm with 200pF cap in parallel.