I am currently developing an automated measurement system for measuring inductance and resistor values of coils.
As hardware I use a PXI-4072 DMM and PXI-2501 SWITCH.
I am using Labview7.1 and my program is based on the example of niSwitch DMM switch handshaking for what concerns the measurements.
However, the readings I get for measuring 10 coils show a reasonable ofset.
For resistance readings, I notice an ofset of about 3.53kOhm and when I measure an open loop, the DMM returns a value of 8.6kOhm
Also, when I measure an inductance of 280mH, the DMM returns a value of about 400mH.
What can be the reason for this corrupt values? I tried to compensate for it by modifying the configuration values of the DMM, but this did not solve anything.
(9,112 Views)
Hi Lieven,
Check this link, it can be useful for you.
http://sine.ni.com/manuals/main?searchsubmit=y&q=DMM+measurement
Regards
Julien Roland - District Sales Manager
NI Belgium - Technical Support
Don't forget to rate a good answer
(9,023 Views)
Thanks for the response Julien.
However, I already found out that the on resistance of the switch causes these offset measurements. In fact, the switch is not very suited for DMM measurements.  I am a bit astonished that the sales representative of NI proposed me this switch, because
it is marked in the DMM manual to use a switch with a very small on-resistance.
Do you know whether it is possible to swap it?
I doubt this is possible, but I would find it the right thing to do for NI.
Lieven
(9,018 Views)
Lieven,
For checking if there is a possibility to swap your board. Please call us at this number 02/757.00.20 (Belgium franch and Dutch).
(9,014 Views)
Thanks Julien,
I just called NI and it is possible to replace the switch....
NI has got itself one pleased customer more Smiley Happy