PXIe-4150 vs. PXIe-4139 Pre-Sales Question, Testing Inrush Current of High Capacitance (10mF...30mF) Loads.

Solved!
Go to solution

Hello, apologize in advance if this has been asked, I didn't see it.

 

This is a pre-sales question for an automated system intended to replace bench tests. The products being tested have a relatively large input capacitance (10mF...30mF) on the power input and relatively low steady-state current draw. There is little no current limiting series resistance, mainly limited by power supply foldback. In some cases power can be hard-switched on (e.g. via electromechanical relays / EMR or field-effect transistor / FET). In other words a delta T of perhaps 200us (200 microseconds) to 1ms (1,000 microseconds). The power being switched in may be 5V, 12V, or even as high as 30V. Using a basic I = C*dV/dt equation it's clear the theoretical inrush current blows up to an unrealistic number (because of course the source and path will have some resistance, the capacitors will have some ESR, etc.)

 

I = C * dV / dt

 = 0.03 F * 30 V / 200E-6 s

 = 4,500 Apk

 

Realistically, in some cases we have measured 20Apk to even 70Apk, which gives one an idea of how much foldback / equivalent series resistance is present in the system (I'll spare you that math 🙂 ).

 

The existing equipment we're attempting to replace for this (and other) measurements is a Tek scope with a hall effect current probe (model numbers aren't really important). We've found some clever ways to get around overloading the equipment, but essentially we need two different measurements:

  • one inrush measurement at fast sampling rates and low Amps/div setting on the scope, and 
  • one long-term measurement at slow sampling rates and high Amps/div setting on the scope.

 

Going forward with automated equipment, I think it's reasonable we don't need to switch as hard (test equipment can reasonably limit inrush), but I need to have a comparative uncertainty analysis to be sure we can still test and reasonably detect an off-nominal input capacitance condition. 

 

For example, knowing an inrush current is 10Apk for the system, should be fine. Let's say capacitor limits are +/-1%, we have 100x better basic accuracy within those limits using a 16-bit DAQ sampling at 1MS/s across a sense resistor. (No, we are not going to use an LCR meter for this application, I'll spare you the details.)

 

I am planning to have NI switches in the circuit to switch among multiple units under test, probably an EMR for this high inrush test, but I'd rather put the work on the supply or our own PMOS FET rather than wear out the EMR switch by hot-switching. As such, I'm not satisfied with the detail in the 4150 and 4139 specifications and I'm asking for additional info. For more background, 

  • In the 4150 configuration an example of the system is something like PX6771950, and
  • In the 4139 configuration an example of the system is something like PX6772373 (see PXI Advisor.)

 

PXIe-4150 questions:

  1. Is an external supply (AUX POWER IN) really required to achieve performance in Figure 1 of the specs?
    • My initial reading made me think "no" because I saw no such note in the specs or datasheet, however...
    • In the PXI Advisor it specifically says so (see image below)
    • If required, the external supply is crazy expensive (almost as much, or as much) as the 4150 itself... just an observation. (I guess that was passive-aggressive for any Sales people reading.) I'm tempted to just plug in a $200 benchtop supply instead.
  2. More driver question than anything - see table 11. I would like to just get a 30V step with maximum output current capability for >=5ms. What is the current range (is it the max 10A) if I configure a 30V voltage-mode step?
    • If so, is there any idea how much foldback (excess current for a brief <=5ms period) there would be? (10Apk for 10A range, 20Apk for 10A range, etc.)
    • My planned method is to use a sense resistor and Multifunction DAQ to determine current at turn-on.
    • Option 1 Voltage mode: As long as I know that output current capability max of the 4150 is 10A or so, I may be able to detect deviations in peak current and reliably assume the deviations is due to the product, not the test equipment.
    • Option 2 Current mode: Alternatively, if I can't get enough current (10A or so) in voltage mode, perhaps I need to configure current mode and do some math to estimate how it will affect voltage on each individual product. Seems riskier... 
  3. Are there any max load capacitance needs of the 4150? I only see a note about 1uF on a noise specification.

    PXIe-4139 questions
  4. Figure 1 of the specifications gives a fairly clear picture of what this instrument can do at 10A. It shows 10Apk for 1ms...10ms at 30V. Is there any way to exceed (in the driver) 10Apk for shorter (lets say 1ms) pulses?
  5. More a driver question than anything - see Figure 1 - is the 1ms...10ms driver limited somehow (e.g. will throw an error if exceeded) or is it just a hardware limitation that shows up without driver error? I'm pretty sure it's the later, because if it's the former, I don't know a method to avoid errors at 10Apk, various supply voltages, other than "guess and check" each pulse width. 
    • If hardware as I suspect, could I get any characterization / finer granularity on pulse widths (e.g. short to ground) for these cases? See also image below, mark-up of Figure 1
      • 10Apk 3.3V output
      • 10Apk 5.0V output
      • 10Apk 12V output
      • 10Apk 20V output
      • 10Apk 30V output 
  6. Same as 4150 question - are there any max load capacitance spec of the 4139? I see no notes.
  7. Per note 11, I don't see the article "Examples of Determining Extended Range Pulse Parameters and Optimizing Slew Rate using NI SourceAdapt" when I search for it on ni.com or on another search engine.

 

andrew_psemc_0-1733856609537.png

 

andrew_psemc_1-1733861071300.png

 

Thanks, 

Andrew

0 Kudos
Message 1 of 11
(548 Views)

Hi, Andrew.  You have a lot of helpful detail in your question, but I’m confused about the big picture.

 

What exactly do you need your automated test system to do?  On the one hand it sounds like you want it to power your DUTs and measure their current draw, but on the other you say you’re replacing a scope plus current probe, which obviously can’t supply any power to the DUT.  Do you want the system to supply power, or just measure current from a separate power supply?

 

Assuming you need the system to supply power, how much power do you need to supply?  You mention measuring inrush with “low Amps/div” and long-term current with “high Amps/div” which sounds backwards to me.  What is the continuous current draw of your DUT?  If the power requirement is low enough, you should be able to use a PXIe-4139.

 

I’m also unclear on if you actually want to measure a high inrush or if you just want to be able to infer the bulk capacitance from the inrush you see, in which case having the SMU limit the inrush shouldn’t be a problem.

 

Then you throw in measuring the current with a DAQ board and a shunt when you’re already sourcing with an SMU, which can directly measure the current it’s supplying.

 

So you can see I’m a bit confused.  I will say that extended range pulsing on the 4139 probably doesn’t work like you think it does.  Exceeding the standard power limit is done by generating a pulse in the driver, which checks the requested V and I setpoint and limits vs. the duration of the requested pulse.  So it is driver-limited.

 

I don’t know that much about the PXIe-4150, but I do know it requires an external power supply of some sort.

 

Anyway, if you can clear up my confusion a bit, I may be able to be a bit more helpful.  

0 Kudos
Message 2 of 11
(515 Views)

"What exactly do you need your automated test system to do?"

Both supply power and provide measurements.

 

The current system and procedure correlates 3-4 independent benchtop instruments (voltage, current probes, scope, and power supply) that I believe can all be done with one SMU.

 

"Then you throw in measuring the current with a DAQ board and a shunt when you’re already sourcing with an SMU, which can directly measure the current it’s supplying."

Assuming the SMU can't measure the inrush current in the way I expect, maybe I'll need to add a DAQ and a sense resistor, but I doubt that will be the case (more details below.)

 

"Assuming you need the system to supply power, how much power do you need to supply?"

Depends on the time frame, obviously, and I'm suggesting we need to soft-switch for production test. To do so, I'm suggesting we break apart the tests into two tests - 1 for inrush and 2 for during steady-state - using an SMU such as the PXIe-4139.

 

Test 1 - set current pulse value and duration, measure dV/dt, calculate capacitance. Because of the 10Apk limitation of the PXIe-4139, this doesn't turn on the DUT, it only characterizes the input capacitance. I argue this is OK and meets the intention of what we were already doing.

 

Test 2 - set output voltage and wait for DUT to fully turn on, then measure steady-state current.

 

Assuming I lose this argument (break apart the existing test into two separate tests) I think I'd need to use something like the PXIe-4150 supply instead of the PXie-4139 SMU.

 

"You mention measuring inrush with “low Amps/div” and long-term current with “high Amps/div” which sounds backwards to me."

Yikes, I said that?! Sorry about that! You're correct.

 

"What is the continuous current draw of your DUT?"

<100mA

 

"Exceeding the standard power limit is done by generating a pulse in the driver, which checks the requested V and I setpoint and limits vs. the duration of the requested pulse.  So it is driver-limited"

I think this answers my question #4, which I pretty much expected, so I'm more preferring the additional information I requested in queestion #5 (specifically, I'm looking for better resolution on the number of ms allowable for each of the 10Apk pulses I mentioned.)

 

"I don’t know that much about the PXIe-4150, but I do know it requires an external power supply of some sort."

Hmm... ok, that's disappointing. Both because the external supply pretty much doubles the cost, and also because it's not stated in the specifications very clearly (e.g. the 4150 graphs.)

 

So I'm trying to keep focused on the high-level - does this make more sense to align with my previous questions?

0 Kudos
Message 3 of 11
(472 Views)

First, I'd like to make sure you realize there are two versions of the 4139: the original 20W version and a newer 40W version.  Their pulsing specs are different, and the pulsing graphs you pulled are for the 40W version.  The 20W version didn't merit a graph, I guess, since its specs are pretty straightforward.  Let's start with that one.

ChrisR_0-1734996525820.png

That 0.2 J energy spec is the one you need.  It tells you how much energy you can get from a single pulse.  So taking your 20V example, 20V x 10A = 200W.  0.2J / 200W = 1 msec.  Similar math for all the other voltages.

The "maximum cycle average power" spec is stating that if you're pulsing 200W, the downtime after your pulse has to be long enough to keep the average output power to 10 W.  If the output power during the off time is 0, that means you'll have to wait 19 times as long after the pulse as the pulse was on.  e.g. 1 msec on, 19 msec off.  The "maximum duty cycle" spec means that 5% is the maximum allowed duty cycle for an extended-range pulse even when the power levels are lower.

 

If you're not familiar with how pulsing is specified in NI-DCPower, I suggest you take a look at https://www.ni.com/docs/en-US/bundle/ni-dcpower/page/pulsing.html, which might clarify some terminology.  One thing that might not be obvious is that an SMU can do pulsing without using the "pulsing" API as long as the requested levels and limits fit within the DC power capabilities of the device.  The pulsing API is primarily to get access to pulsing beyond the DC limits.

 

The specs also contain the information you requested for the 40W 4139.  Look for the piece of the specs that starts like this:

ChrisR_1-1734997116878.png

I won't try to work through the equations here, but I think you should be able to figure them out.

 

One thing you might consider trying is to create a simulated 4139 and see what the software allows you to do.  I'm no expert on how to set that up, but it should be pretty straightforward.  It used to be done in MAX, but now I think there's a newer tool to do it.

 

I now understand better what you're trying to do, but I'm still a bit confused, partly because of comments like this: "Because of the 10Apk limitation of the PXIe-4139, this doesn't tu