PCI4070 NI-DAQmx - creating virtual channel - no supported devices found

Solved!
Go to solution

I use PCI-4070 to measure the Voltage and Current. The self-test and test panel display well, also display well on Windows device manage. But when I use NI-DAQ>>Aquire Signals>>Analog Input>>Voltageit , it said no supported devices found.

I am using LabVIEW 2016 version. Could anyone tell me why?

Download All
0 Kudos
Message 1 of 8
(6,235 Views)

The PCI-4070 is programmed using NI-DMM, not NI-DAQmx.

I would recommend checking out the NI-DMM examples in Help->Example Finder to get started.

Tobias
Principal Software Engineer
Driver Software
National Instruments
0 Kudos
Message 2 of 8
(6,224 Views)

Thank you! You give me the right direction. I cannot find the NI-DMM examples in my LabVIEW following the guide form :

To browse the NI-DMM examples available in LabVIEW, launch LabVIEW, click Help»Find Examples, and navigate to Hardware Input and Output»Modular Instruments»NI-DMM (Digital Multimeters).

 

and as I try to follow the tutorial in link

http://zone.ni.com/reference/en-XX/help/370384N-01/dmm/lv_tutorial2/

  1. Open an existing or new LabVIEW VI.
  2. From the Function Palette (the default location), locate the NI-DMM VIs at All Functions»NI Measurements»NI-DMM.
  3. Select the VIs that you want to use and drop them on the block diagram to build your application.

But I find nothing about the measurement block in Functions.

I checked that the NI-DMM driver is installed.

Are there any problems? 

Download All
0 Kudos
Message 3 of 8
(6,205 Views)

What version of LabVIEW are you using? What version of NI-DMM? Did you install NI-DMM after LabVIEW?


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 4 of 8
(6,200 Views)

PS, now my driver is version 17.1 and my LabVIEW version is 2016, is that the problem?

0 Kudos
Message 5 of 8
(6,199 Views)

I have said on the first article. And the screen of the MAX is attached on it. 

LabVIEW 2016

PCI-4070

NI-DMM version 17.1

0 Kudos
Message 6 of 8
(6,198 Views)
Solution
Accepted by topic author NewFifty

@NewFifty wrote:

PS, now my driver is version 17.1 and my LabVIEW version is 2016, is that the problem?


That should be fine.  Try reinstalling NI-DMM to make sure the LabVIEW support is installed.


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 7 of 8
(6,191 Views)

It works for the driver 16.0