05-05-2023 09:26 AM
I am trying to set the 4071 and 4081 DMM's input resistance via the DMM property node. I have gathered from other posts that the range of possible settings (1M, 10M, >10G) is limited by a combo of Auto Range, Range, and (possibly) Auto Zero settings. I am unable to get the device to enter 1M mode. Any thoughts on what settings are required to allow the devices to enter 1M input resistance mode?
Solved! Go to Solution.
05-05-2023 10:23 PM
These details are typically documented in the DMM help file or the user manual.
05-06-2023 12:31 PM - edited 05-06-2023 12:31 PM
Available input resistance settings are listed in the product specifications. 1M is not an option for the 4071 and 4081. You can do 1M in digitizer mode on the 4070/72/80/82.
05-09-2023 02:01 PM
Thank you, @ChrisR.!