10-27-2009 07:46 AM
You can have one master and up to 15 slaves. However, if all of the responses have the "<" to start, then your current technique will not work.
Can you post your current worksheet file or make a small example with just the RS232 modules and the Action?
The next step is to contact your reseller directly. This is not an efficient way to assist you!
10-27-2009 08:06 AM
Attached you find an example of the problem.
Each button sends a different command to RS232.
This is only possible if you use 'restart on timeout' in options of the input module,
otherwise it will execute only the first pushbutton after programstart.
Regards
Andreas
10-27-2009 09:12 AM
It is possible that you have not accounted for all of the characters.
Try setting the last command for 1x 5a $06\r
Let me know if that works.
10-27-2009 09:20 AM
Yes, now it works as expected !!
Thank you very much !
Kind regards
Andreas Graff