Counter/Timer
Finite Pulse Train Generation - Exceeding Counter Limits
I have an application where I am using 1 counter as a master clock for 4 other counters, in order to generate 4 synchronized signals.  The idea was that by controlling the master clock, I can easily change the pulse rate of the 4 signals, yet keep them in perfect sync with each other.  I use this in both Continuous and Finite modes.
The issue is that under one condition, the counter is returning an error when I call START:
ERR: -200305
Possible reason(s):
Desired finite pulse train generation is not possible.
Change the number of samples to be generated, increase the rate of the pulse train, or choose a different timebase source. Refer to the documentation for more details.
Property: CO.Pulse.Freq
  Corresponding Value: 230.0
Property: CO.Pulse.DutyCyc
  Corresponding Value: 500.0e-3
Property: CO.Pulse.Freq.InitialDelay
  Corresponding Value: 0.0
Property: CO.CtrTimebaseSrc
  Corresponding Value: /Digital/80MHzTimebase
Property: SampQuant.SampPerChan
  Corresponding Value: 20000000 (That's 20,000,000)
Channel Name: Digital/ctr4
Task Name: _unnamedTask<6>
Any idea how I can accomplish this?  Note:  Essentially this is a ~24 hour task (20M/230 = 89,000 sec).  Do I need to set it up to do, say, 2M pulses and then retrigger?  Can I set that retrigger up for a fixed 10X?
Also note:  In 99% of the customer's use cases (which are much shorter), this scheme works.  This is a new one they surprised me with, otherwise I would have tested it.  So any solution should be a general solution that will work for 230Hz up to 16,000 Hz, and possibly up to 100K samples.
Thanks!
(2,943 Views)
What's your hw device?  Lacking any other info, I'll start by supposing something like a 6602 counter/timer board that has a 80 MHz max timebase and has more than 4 counters onboard.
1. On that device, a finite pulse train will (invisibly) require the use of 2 counters, both the one you program for the output and the adjacent one.  The adjacent "helper" counter generates a single pulse which acts like a pause trigger for the output counter.  It's the duration of the helper counter pulse that controls the # of pulses generated by the output counter.
   So you need to be sure the adjacent counter is available.  Based on the error message you're seeing, I'm pretty sure this isn't your problem, but figured I should mention it just in case the adjacent counter is only available by *coincidence* rather than by explicit choice.
2. I would expect (but can't say for sure) that DAQmx would try to use the same timebase for both the output counter and the helper counter.  Here that's 80 MHz.
3. An 80 MHz timebase will overflow a 32-bit counter in less than 54 seconds (2^32 / 80e6).  Have you seen success with use cases where the master clock runs for a minute or more?
4. It's also possible that DAQmx is smart enough to use the slowest onboard timebase, 100 kHz, for the helper counter.  This will overflow a 32-bit counter in just under 12 hours.  Perhaps *this* is the breaking point between successful and unsuccessful use cases?
5. I'll suppose for now that the breaking point is ~12 hours.  So now the next question is: when a use case lasts for 12+ hours, how *exact* must the # pulses be?  In other words, is it conceivable you could live with running your master clock in Continuous mode, and then use software timing to decide when to stop it?   You'd need to allow for several seconds worth of uncertainty in your pulse generation duration if you choose this approach.  Some of it due to normal software timing uncertainty, most of it due to timing skew between the counter timebase and the real-time clock used to make the software decision.
6. If you *must* have a more nearly exact # of pulses over the course of 12+ hours, you'll need a more complicated approach.  Instead of using 2 counters to generate a finite pulsetrain from your master clock you'd need 3.  *And* you'll need to manage the finite aspect of pulse generation more manually rather than letting DAQmx do it automatically.
7.  You'll configure output counter A for continuous pulse generation and also configure it to be pause-triggered by the single pulse output of "helper" counter B.  In turn, counter B will need to derive its pulse timing from a timebase that can be slower than 100 kHz.  Let's use counter C to generate that timebase, and when you need 1+ days worth of pulses, let's set it to generate pulses at 5 kHz (that'd give you ~10 days long single pulse from counter B meaning ~10 days of output counter A pulses).
8.  Counter A can continue to use the internal 80 MHz timebase for precise output pulse timing.  Counter C should be in continuous generation mode and should be started first.
Next you should *commit* Counter B to force its output signal to be in its defined "idle state".  (Counter A needs to be configured to pause when Counter B is in idle state.)  Next start Counter A, and then finally start Counter B.
So, you can see why it'll be nice if approximate software timing can be "good enough" for use cases >12 hours.   Smiley Wink  All this complicated stuff can be used for the general case if necessary, but it's gonna take some work to wring out all the debugging details.  In my opinion, the best-case scenario is if your current breaking point is ~12 hours, and longer use cases are not so strict about exact #'s of pulses.   Your use of a master clock will *always* guarantee that the other 4 *meaningful* counters start and stop in sync.
-Kevin P
(2,927 Views)
Now THAT is a thorough reply.  😉
Yes, it's a 6602, and it works fine with lower counts or faster ticking
Do you know what property would reveal which counter is being used?  I am explicitly using counters 0-3 for the signals, and explicitly using 4 for this... so it would have to be ctr5.  But how to verify?  Maybe not worth the trouble, but nice to know.
Absolutely.  I have run it for a couple hours.  Playing with the settings, With 1000Hz it tops out at 42,950,000 steps;  just under 12 hours
Again... would that be ctr5?  I can check it.
So yeah, it looks like 12 hours; And you are correct... right now the solution is a software timed loop.  And this is acceptable for this situation... however I'd like to find a generalized HW solution, since now I have to figure out how to let the operator know when to use either code base.  Ugh.
OK, I'll code that out to wrap my head around it.  Sounds straighforward.  thank you!
-Kevin P
(2,920 Views)
The idea of adjacent (or "paired") counters is fixed at the hardware level for the 6602.  The pairs are (0,1), (2,3), (4,5), or (6,7).  I don't know of a property for querying it.
If you manage the helper counter explicitly, Counter 4 can be paused by the output of any other available counter (5, 6, or 7), you don't *have* to use Counter 5.  I probably would anyway though, no good reason *not* to.
-Kevin P
(2,915 Views)
