Your GPIB card is telling you that the DTAS line is active. This would normally indicate that some other device on the GPIB bus is acting as the bus controller.
You might want to try and talk to the device directly. Open Measurement and Automation Explorer. On the left side expand Devices and Interfaces. Expand GPIB0 (assuming your card is installed as the default name). If there are not any instruments shown, then Scan for Instruments. You should see a list of instruments. If you right click on the instrument and select Communicate with Instrument, you can try sending commands directly to the instrument. You can also see the status of the GPIB interface lines in the window.