When you use Scan for Instruments in MAX, VISA isn't used. Only the low level GPIB functions are used to find what is connected and send the *IDN? query.
You say they are not the same GPIB controllers? Are they both from NI?
Have you tried swapping GPIB cables or tried swapping instruments? It seems to me that there may be an actual hardware problem somewhere.