05-29-2011 11:51 AM
I would like to interace a laptop with a Keithley 195A (Digital Multimeter).
The only available hardware interface is the GPIB.
On a laptop the available interface is USB.
Than a Prologix USB-GPIB Controller 6.0 device is what should do the job.
Attached there are a sample program from Prologix, a (hopefully) working program for the DMM from Kithley, and what I did (a simple copy and paste).
Result: after running the command ++addr 16, with the program from Prologix (just to start the communication.. I know, I will optimise the program) the program I realised do change the setting of the DMM. Say, I set a Ohm measuring, and the reles switch, but not value is read from the instrument.
What do I miss, or where is the error?
Many thanks
Solved! Go to Solution.
05-29-2011 04:54 PM - edited 05-29-2011 04:56 PM
Solved! ![]()
The following command was missing:
++auto 1 //Turn on read-after-write and address instrument to talk
Let's now optimise the program! ^^
05-29-2011 07:57 PM - edited 05-29-2011 08:03 PM
Here there is the final version! ![]()
05-31-2011 06:59 AM
04-02-2012 10:27 AM
I've just got Prologix USB-GPIB adapter and I cant make LV and MAX to find it . Could you help with this?
I have Windows 7 . I have installed 2.08.14 WHQL driver for Prologix and NI 488 and VISA drivers.
MAX can not find any device. although, Prologix GPIB Configurator communicate with my GPIB instruments OK.
04-02-2012 10:33 AM
Why would you think MAX should find it? The only thing MAX or LabVIEW is going to find is a new serial port. Installing NI-488 was a waste of time since you are not using an NI 488 controller.
05-30-2017 05:31 AM
Hello mate!
I aprreciate your answer with your definitive program.
I was doing a program for the Impedance Analyzer HP 4192A and it helped a lot.