Signal Conditioning
reading wrong voltage scc-68
I am using LabView Signal Express that came with my SCC-68.  I have several voltage sensors connected:
RSE at Ai1    A+=Pin33, A-=Pin 67
RSE at Ai2    A+=Pin65, A-=Pin 32
Diff at Ai3     A+=Pin30, A-=Pin 63
Diff at Ai4     A+=Pin28, A-=Pin 61
RSE at Ai5    A+=Pin60, A-=Pin 27
My problem is the Ai3 signal.  I am trying to read a mV representation of the current of a motor.  I hooked up a Fluke clamp current meter which outputs 1mV per 1 Amp sensed.  I have another multimeter connected directly to the same points where the signal is connected to the SCC-68.  On the meter I read 0 mV and when I increase motor current up to 11 Amps ... I read ~11mV.  HOWEVER, I read around 15mV on the Ai3 input the entire time.  All other inputs are working fine.  I even tried to move this input to another input (Ai6) but got the same results.
The only thing I can think of is that all the other inputs are referenced in some way to one power supply where the current one runs off it's own battery.  However I tried to tie the return side to the same ground to make it a common reference but it didn't fix the problem.   I also experimented with changing it to a RSE and so forth.
Any ideas?  I am truly new to this because we just got this equipment.
So to review:
- getting incorrect reading from 1 of 5 signals
- reading voltage directly at connection to SCC-68 with a multimeter and not reading the same as the SCC-68 is
- that one is a meter that runs off it's own 9v battery
- all others run off a common 5v power supply
HELP please!
(6,251 Views)
Can you provide a schematic of your application? Like you mentioned it sounds like a signal reference issue.
Glenn
Glenn
(6,235 Views)
