Signal Conditioning

cancel
Showing results for 
Search instead for 
Did you mean: 

is there protection on the excitation channels of the scxi-1540

Solved!
Go to solution

Hi All,

 

Is there any protection built into the circuitry of the SCXI-1540 so that if the primary winding of an LVDT short circuits then the excitation from the 1540 will not be  destroyed.

 

kind regards

Frank Greenslade

Christchurch

New Zealand

0 Kudos
Message 1 of 2
(6,298 Views)
Solution
Accepted by topic author FRG

Hi Frank,

 

The SCXI-1540 is supplying the excitation, thus the same amount of excitation is provided by the module even if the LVDT shorts.  No additional current should be drawn.  Hopefully this helps!

 

 

Regards,


h_baker
National Instruments
Applications Engineer
0 Kudos
Message 2 of 2