Signal Conditioning
configuration check request
we need test 10 pieces of thermocouples ,3 LVDT signals and 3 strain guage at one time.we both want the configuration of the signal conditioning and DAQ device meet our requirements and the low cost,so we select the following configuration:
SCXI-1000
SCXI-1520
SCXI-1330(instead of SCXI-1314)
SCXI-1540
SCXI-1330(instead of SCXI-1315)
SCXI-1102
SCXI-1330(instead of SCXI-1300)
PCI-6023E(12 bit is enough)
SSH6868
SCXI-1360,SCXI-1361
we hope these selections can make a whole DAQ system,could you check the configuration and please tell us if there is existing any mistake,and welcom better configuration. thanks
(3,626 Views)
Instead of the SSH6868, you will need a shielded cable assembly to connect your 60xxE (E-series card) to the SCXI chassis. You will require the SCXI-1349 cable assembly. It has two parts 1)SH6868 cable and 2) an adapter to connect to the SCXI backplane.
As for the choice of the DAQ card,you may find the low cost approach severly limits the system bandwidth and overall accuracy. The low cost E-series boards only have 512 samples of onboard memory, as opposed to 8192 for the regular E-series. This makes controlling buffered analog input more difficult. In addition, you need to understand that the rated sampling rate for the board is based upon a single data channel. As you have 16 channels, your "best" sampling rate would be approximately 12.5 KHz per channel. The actual rate will be less. You have made no mention of the type of data you want to acquire (steady-state, dynamic, time domain, frequency domain), so, I can not give any better insight into the "best" sampling rate you could achieve.
Now, considering that you are mixing thermocouple (TC) readings with other inputs on the same DAQ call, you may find the 12-bit selection aggravates your data quality. When you acquire data using a multiplexer, the quoted accuracy for the device assumes that all of the voltage levels on the various channels are very near the same value. This allows a relatively uniform settling time for the input signal. It is quite possible that the TC inputs will be on the order of 2-10 mV while the other signals could range as high as 0.5 volts. Such a difference in input levels can lead to settling errors that are not obvious to the user, but can cause gross errors in the measured data levels. If you set the input range (only 4 for the low cost e-series) to +0.5/-0.5 volts, your resolution for the TCs will be about +/- 0.24 mV or about 6-10 degrees of error. (ouch!!). So you will be required to set different input ranges for the different input channels (0.05/-0.05 for the TC and +5/-5 or +.5/-.5 for the remaining channels. These multiple range calls can further limit your bandwidth by slowing down the "round-robin" interval between the data range changes. I know that your goal is low cost; however, you may find that the data quality is unacceptable. You will want to consider, at the very least, upgrading to the 6034E board. It is a 16-bit device (still only four input ranges); however, if you configure the board for only 1 input range, +0.5/-0.5 volts, your bandwith will improve and the TC resolution will improve by a factor of 16 times to approximately 0.015 mV (under 1 degree).
(3,626 Views)
thanks a lot for the detail explanation,that make me re-think the configuration.the type we want to acquire isn't very dynamic,for instance,the temperature measured by the thermocouple varies from 0 to 120 degree in nearly 3 minutes, the maximum frequency is nearly 100 HZ. the similar parameter as to the strain gauge(full bridge) and the LVDT.
yes, we think we need measure the 16 analog inputs(10 from thermocouple,3 from strain gauge and 3 from LVDT) together sometimes,at least the signal of TC and the signal of strain gauge should be measured together. under such situation, do you think what kind of input range do we set is best for us?
now i rerange the configuration for our requirements, shown below, how do you think this, welcom any better recommendation:
SCXI-1000;
SCXI-1102;
SCXI-1540;
SCXI-1520;
PCI-6034;
SCXI-1349;
SCXI-1315(3 pieces)
SCXI-1360,SCXI-1361
thank you!
(3,626 Views)
