02-16-2007 07:44 AM
02-17-2007 01:01 AM
I can only answer your Q1
It is not possible to have diferent sampling rates for each module
Since all your Modules/channels are multiplexed to the DAQ card and the ADC on the DAQ card is shared between all modules/channels, you cannot have different sampling rate for each module/channel
What you could do is to sample all modules/channels at highest required/ possible rate and decimate those channels in which you want a lower rate
02-17-2007 09:21 AM
02-19-2007
09:41 AM
- last edited on
09-10-2026
12:14 PM
by
Content Cleaner
Hello Kater.
The SCXI-1112 sample rate per channel is directly dependant on the number of channels being sampled at the same time.
Please refer to the following knowledge base for the ceiling on the sample rate for your particular applicaiton:
Recommended NI Hardware for Thermocouple Measurements
http://digital.ni.com/public.nsf/websearch/F0196F5B6CD8F60286256D4B00728B78?OpenDocument
As far as a good example for thermocouple measurements with SCXI modules and VB.net, please navigate to the following directory:
C:\Program Files\National Instruments\MeasurementStudioVS2003\DotNET\Examples\DAQmx\Analog In\Measure Temperature\ContAcqThrmstrSmples_IntClkSCXI\Vb
where the C drive is your harddrive. This gives you a great example of acquiring measurements with SCXI modules. To add another module, you will need to create another task by following the procedure followed in the above example.
Let us know if there is anything else that we can do for you!
Brian F
Message Edited by Brian F. on 02-19-2007 09:46 AM
02-19-2007 10:20 AM
02-20-2007
03:11 PM
- last edited on
09-10-2026
12:14 PM
by
Content Cleaner
Hello.
It is ok to connect all 4 modules to one DAQ card. As long as you connect one of the modules to the card, all modules will be able to communicate with the DAQ card via the cable and the SCXI backplane. The number of samples achievable is highly dependent on the channel count on each of the modules. As some background, the SCXI-1125 modules have a 333 kS/s maximum, random scanning sampling rate. The SCXI-1112 module has a maximum rate of 100kS/s / (2*the number of channels). All of these samples will be multiplexed by the PXI-6280, which has a maximum sampling rate of 500 kS/s. Thus, the PXI-6280 might be your bottleneck, limiting your acquisition rate at 500 kS/s. For most applications, this should be more than adequate.
In terms of creating multiple tasks in DAQmx, this is allowable at the same time. One task for each set of parameters should suffice. For each task, you will follow the basic procedure for creating one task as outlined in the example that I pointed you to in my last post. For more detailed information on this process, please refer to the following Developer Zone article at:
Using NI-DAQmx in Text Based Programming Environments
https://www.ni.com/en/support/documentation/supplemental/21/using-ni-daqmx-in-text-based-programming...
Good luck with your application!
Brian F