Signal Conditioning
SCXI-1520 AND 700 OHM LOAD CELL?
Will the SCXI-1520 work with a 700 ohm full bridge load cell? I'm a load cell mfg proposing a 700 ohm load cell solution to a customer. He says he doesn't think the 700 ohm cell will work with his 1520. The load cell will can take a maximum 20 VDC excitation and outputs 2 mV/V.
Message 1 of 8
(5,016 Views)
The 1314 should have no problems with a 700 ohm full bridge load cell. The module can supply 0-10V of excitation to the load cell when using an SCXI-1314 terminal block attached to the 1520. Since this is a full bridge sensor, there isn�t even any need to change any bridge completion resistors inside the 1314, so it should just be a matter of wiring up the inputs and excitation outputs to the sensor.
I hope this helps! Please let me know if there is anything I can help with or clarify.
Ken S.
Applications Engineering
Message 2 of 8
(5,014 Views)
I have a similar problem with 1520 and SCXI 1314 terminal block. I connected one load cell of max capacity 200 kg(f) but not able to get any signal. What ever signal it shows is some drifting fluctuating signals. I configured it using both analog voltage input as well as analogue strain input. Still does not work (using MAX).
Message 3 of 8
(4,772 Views)
What kind of fluctuations are you seeing? Is the excitation coming from the 1314 correct? Have you tried using a different module?
Abhinav T.
Abhinav T.
National Instruments India
LabVIEW Introduction Course - Six Hours
Getting Started with NI-DAQmx
Measurement Fundamentals
Message 4 of 8
(4,746 Views)
I am also using SCXI -1315 for LVDT. It is working properly. The only problem is with 1314 block which gives some random signals only.
Message 5 of 8
(4,710 Views)
Can you verify that the voltage coming out of the thermocouple is correct? You can do this by measuring across the thermocouple using a hand held multimeter. You may also want to input a known voltage into the module (a simple battery would suffice) and see if this produces the correct voltage in Measurement and Automation Explorer.
Abhinav T.
Abhinav T.
Message 6 of 8
(4,690 Views)
I have verified that the load cell is working properly using a mutimeter before posting to this thread. What it gives is some fluctuating signals with ever increasing values.
Message 7 of 8
(4,688 Views)
Can you try connecting a known voltage such as a battery to the S+ and S- inputs of the 1314 and see if this voltage is displayed correctly?
Abhinav
Abhinav T.
Message 8 of 8
(4,670 Views)
