Signal Conditioning
1303 current measurement configuration.
What signal source can you use with the 1300 (ie. no bias resistors)?  I am having an issue with ground referenced signal on the 1300.
Message 11 of 17
(2,733 Views)
BME_Genius,
What SCXI module are you using along with the 1300 terminal block?  Also, what type of signal are you trying to measure?
Message 12 of 17
(2,726 Views)
I am using the 1100 module.  The signal is a 0-7.5 Vdc from a CO2 analyzer.  The output impedance is 20kohm.
Message 13 of 17
(2,721 Views)
Have you verified the jumper settings in the 1100 module correspond the the grounding type of the signal source?  A description of the jumpers is available in the 1100 user manual:
https://www.ni.com/docs/en-US/bundle/320637c/resource/320637c.pdf
Message 14 of 17
(2,700 Views)
Yes.  Jumper W1 is in position 1-2.  However, it cannot be a module problem because it only occurs when that signal is attached to the 1300.  This was never a problem before with this signal on 1300 and started to occur intermittently and then would not go away.  The problem signal still is and always was ground referenced.
Message 15 of 17
(2,693 Views)
Interesting...
It seems that a ground-referenced signal shouldn't need any special wiring considerations...
From the manual:
"If you are using a ground-referenced thermocouple, connect the
thermocouple positive and negative leads to CH+ and CH– respectively.
Leave jumper W1 on the SCXI-1100 in the factory-default (parking)
position"
This is how you have if wired, but you're still having his issue, correct?
Also, does this behavior hold regardless of input channel?
Message 16 of 17
(2,670 Views)
Yes, this is how I have it wired and always has been wired on the 1300.  No matter what channel I use on the 1300 I see the same problem.  For some reason, it works fine on the 1303.  That gave me the idea to tie the negative terminal to ground via a 10kohm resistor.  This fixes the problem on the 1300...but WHY being it is ground referenced????
Message 17 of 17
(2,661 Views)
